Hello,
Currently I’m working on a phantom project that mimics mouse and human brain metabolites. For metabolite quantification, I have been utilizing the 7T brain basis sets available on the LCModel website to estimate metabolite concentrations in my phantom data. However, I have encountered inconsistencies in the fitting results across various ppm ranges, which suggests that the current basis sets may not be fully optimized for my specific phantom composition.
It would be greatly appreciated if you anyone could share any basis sets suitable for phantom metabolite estimation, as this would be highly beneficial prior to proceeding with the simulation of our own custom basis sets.
We have acquired spectra for individual metabolites as well as a combined phantom spectrum at TE = 18 ms using a 7T scanner, though we are capable of acquiring data at different echo times if necessary.
Any advice, guidance, or available scripts/code you could share for simulating phantom spectroscopy would be extremely valuable and much appreciated.
The issue you are facing probably isn’t to do with suitability of the basis set for a phantom, but rather suitability for the specific sequence (and sequence parameters) you are using. Could you tell us the exact scanner, sequence, and sequence parameters (primarily echo time)?
Thanks for the details. You will need just a few more details to make a decent basis set for your data. Namely:
-TE1 and TE2 (the total echo time is divided into TE1 and TE2 such that TE1+TE2 = TE).
-Refocusing RF pulse waveform used.
-Refocusing RF pulse duration.
i guess i’m answering quite late. We are also running some phantoms measurements and quantifications with LCModel on our 9.4T system (Bruker one). We observed that the phantom data are quite narrow thus we slightly lb the spectra before LCModel (not recommended to do but we had to).
try to see how this might help you
best
cristina